Venting
Procedure S800 Beam Line:
Target Section
Panel mate: s800vac.MT2, page 05.
- Close gate valves (I246GV and I255GV) of the main scattering
chamber section using the panel mate (upper left and lower right on page).
- Turn off the turbo pump controller by pushing the stop switches to
the left (the controller is located on the wooden platform under the beam
pipe in S3). The green light should turn off.
- If it is on, turn off the ion gauge. The controller is located
under the beam line, on next to the turbo pump controller.
- Wait for five minutes to let the turbo pumps spin down.
- Close the hand valve between the turbo and roughing pumps.
This valve is located on the far side of the beam line.
- Open hand venting valve on the near side of the cross piece (You
may open as fast as you wish).
- Note that you are venting to air so there is no danger of over
pressurizing the chamber.
- When the pressure is at atmosphere (~760 torr) close the venting
valve.
- Ready to access the beam line.