S800 Troubleshooting

 

The gas handling system triggers a voice alarm

Please follow these steps

 

The HV GUI triggers a voice alarm

Check the read-back value of the voltage displayed above the "HV set" button for the detector that triggered the alarm. In the example shown below, the read-back equals the set voltage of 1000 V for the anodes of both CRDCs. If read and set voltages match and the scalers/spectra for this detector still increment, the alarm was false and shouldn't come back once acknowledged in the alarm monitor.

A tracking PPAC might occasionally trip and will trigger a real alarm. Turn it back on via the "Turn Off/Turn On" button. If it continues tripping, the rate might be to high. The two solutions are to lower the rate until the PPAC stays operational or to lower the PPAC voltage by 5 or 10 V. Please keep in mind that lowering the voltage increases the performance of the detector. If the PPACs cannot be turned on again at reduced rate or lower voltage or if the CRDCs or the ion chamber trigger a real alarm, contact somebody from the S800 group.

 

The NMR GUI beeps

Please follow these steps

 

The NMR GUI "freezes"

The NMR program is connected to the following devices and systems:

•  The NMR module through a terminal server

•  The digital oscilloscope through a terminal server

•  The EPICS system through the caRepeater process launched by the et_wish shell running the program

Any of these connections can eventually be broken, and the program tries to catch the failures and report them to the user. However there are cases when the program “hangs” and need to be manually aborted. The little heart displayed at the top left corner of the oscilloscope display indicates the liveliness of the EPICS loop, as well as the liveliness of the program itself. If the heart stops beating for more than 10 seconds, it probably means the program is frozen. To restart it, simply click on the program's icon again, which will kill the existing process and start a new one. It has been observed that it occasionally takes two to tree tries to restart the NMR application.

In case nothing happens after trying multiple times, the connection to the NMR module is probably locked and needs to be freed. To free it, type the following from any shell window (actual typing show in this font ):

Click again on the starting icon and the program will start.

 

Resetting the S800 alarm monitor in case of "too many alarms"

on uxpc2 start "et_wish" and enter

 

DAQ crashes that require the S800 VME crate to be rebooted

Crashes of the data aquisition related to problems with the bus of the S800 VME crate have been observed and require a reboot of the crate (power cycling). Those crashes happen for the CRDC as well as for the tracking PPAC readout. A typical error message is "Failed to grab busses in crdc2_pads (slot 8)". The power supplies of the gas-filled focal plane detectors are located in this crate and require special attention. The VME crate is the topmost crate in the S800 electronics tower (toplevel of S3, on the left).

 

UXPC2 has to be rebooted

Log in as user s800, password to be obtained from the S800 group

Start the following applications using the icons (preferably on the desktop indicated in paranthesis):

Enable the alarms and log option on the HV Gui (buttons at the bottom).

Barney is only in exceptional cases supposed to be used by the experimenters.

 